Tue Jul 29, 2014 7:33 pm   



Hi all,

I am looking to calibrate 16 bit A/D converters used in several
configurations (DC voltage, AC voltage, and RTD) and was wondering if a
Fluke 5100B 4.5 digit calibrator would be a good choice accuracy-wise?

According to blurbs on the web, these are the specs.

The 5100B AC/DC Calibrator calibrates precision meters that measure AC or
DC Voltage (0-1100V), AC or DC current (200 uA-2A), and Resistance (1 ohm
to 10 Mohms). Accuracies are: DC voltage and resistance (0.005%), DC
current (0.025%) AC Volts (0.05%) and AC current (0.07%).


Thanks.


Guest

Tue Jul 29, 2014 7:33 pm   



In sci.electronics.basics Robert T. <nospam_at_dev.null> wrote:
Quote:
I am looking to calibrate 16 bit A/D converters used in several
configurations (DC voltage, AC voltage, and RTD) and was wondering if
a Fluke 5100B 4.5 digit calibrator would be a good choice accuracy-
wise?


A 16-bit A/D converter has 65,536 steps over its full range. If you
want to have resolution down to one step, you need something that's
accurate to at least (1/65536)*100 or 0.00153%, and preferably more than
that.

If you're willing to accept being three or four steps off, then the
0.005% is OK. It depends a lot on what you're doing with the digitized
data.

Matt Roberds

Robert T.
Guest

Wed Jul 30, 2014 3:50 pm   



On Tue, 29 Jul 2014 14:31:01 +0000 (UTC) mroberds_at_att.net wrote in Message
id: <lr8b72$5m9$1_at_dont-email.me>:

Quote:
In sci.electronics.basics Robert T. <nospam_at_dev.null> wrote:
I am looking to calibrate 16 bit A/D converters used in several
configurations (DC voltage, AC voltage, and RTD) and was wondering if
a Fluke 5100B 4.5 digit calibrator would be a good choice accuracy-
wise?

A 16-bit A/D converter has 65,536 steps over its full range. If you
want to have resolution down to one step, you need something that's
accurate to at least (1/65536)*100 or 0.00153%, and preferably more than
that.

If you're willing to accept being three or four steps off, then the
0.005% is OK. It depends a lot on what you're doing with the digitized
data.


Thanks Matt. I think that I can live with a few counts off.

elektroda.net NewsGroups Forum Index - Electronics Equipment - Fluke 5100B

Ask a question - edaboard.com

Arabic version Bulgarian version Catalan version Czech version Danish version German version Greek version English version Spanish version Finnish version French version Hindi version Croatian version Indonesian version Italian version Hebrew version Japanese version Korean version Lithuanian version Latvian version Dutch version Norwegian version Polish version Portuguese version Romanian version Russian version Slovak version Slovenian version Serbian version Swedish version Tagalog version Ukrainian version Vietnamese version Chinese version Turkish version