** After just a little research - it is not hard to see why connecting a
scope directly to the AC mains is such a bad idea.
The vast majority of analogue and digital scopes have switched, passive
input attenuators using small resistors, trim capacitors and wafer switches
employing several banks. All these components have max voltage ratings that
do not include supply voltages like 240 volts AC.
Possibly the most vulnerable component to high voltages is the tiny trim
capacitor connected * directly across the input * to equalise capacitance
loading for all ranges - it is crucial that a scope's input present a
constant value of C ( usually 15 to 20 pF) in parallel with 1 Mohms on all
ranges so that the trimmer on a 10:1 probe can be adjusted once and left.
Wafer switches are also not rated for high voltages and lack sufficient
insulation and clearances to survive the spike voltages that regularly
appear on the AC supply. Insulation breakdown in the trimmer or switch
would be sudden and bit spectacular when full AC mains power is available.
Many scopes have a max input rating of 400 volts ( DC+AC) peak - which
means they are capable of accepting such peaks on a wave being viewed
without damage, despite the fact the actual peaks may not be shown. It also
means that when AC coupled, the max DC voltage should not exceed 400 volts
as that is the rating of the series input capacitor.
The worst case of ACCEPTABLE input overload is with 400 volts at the input
and the attenuator set to pass the voltage straight through to the FET
preamp. Makers typically fit a series resistor of 100kohms to 470kohms
between the switch and the gate of the first FET to limit current to a few
mA - then add one or more diodes to direct this current safely into the
+/- DC supply rails. These diodes must be low leakage types so not to create
an DC offset at the gate of the FET.
This high value series resistor needs a cap in parallel to pass high
frequencies that would otherwise be lost due stray capacitance to ground at
the gate of the FET - typically a 10nF cap is used and this too must be
rated for at least 400 volts.
In short - whenever scoping the AC supply make damn sure to use a 10:1
probe and be DAMN certain the switch is the right position.
.... Phil
Phil Allison
Guest
Fri Dec 30, 2011 4:05 am
"Phil Allison"
Quote:
In short - whenever scoping the AC supply make damn sure to use a 10:1
probe and be DAMN certain the switch is the right position.
** Just to be clear, the switch is the one on the side of the probe that has
3 positions: x1, X10 & GND.
.... Phil
fungus
Guest
Mon Jan 02, 2012 1:25 pm
On Dec 30 2011, 2:06 am, "Phil Allison" <phi...@tpg.com.au> wrote:
Quote:
In short - whenever scoping the AC supply make damn sure to use a 10:1
probe and be DAMN certain the switch is the right position.
Thanks for that...once again things are more
complex than I imagined.
Bob Myers
Guest
Mon Jan 02, 2012 10:43 pm
On Monday, January 2, 2012 4:25:06 AM UTC-7, fungus wrote:
Quote:
On Dec 30 2011, 2:06 am, "Phil Allison"
wrote:
In short - whenever scoping the AC supply make damn sure to use a 10:1
probe and be DAMN certain the switch is the right position.
Thanks for that...once again things are more
complex than I imagined.
That was good advice in any event; you should ALWAYS be in the habit of spending a good deal more time setting up the measurement than actually making it. "Grab a probe and stick it on" is a recipe for getting bad measurements and/or frying lots of nice, expensive equipment.
Bob M.
Phil Allison
Guest
Tue Jan 03, 2012 12:23 am
"Bob Myers"
fungus wrote:
"Phil Allison"
Quote:
wrote:
In short - whenever scoping the AC supply make damn sure to use a 10:1
probe and be DAMN certain the switch is the right position.
Thanks for that...once again things are more
complex than I imagined.
That was good advice in any event; you should ALWAYS be in the habit of
spending a good deal more time setting up the measurement than actually
making it. "Grab a probe and stick it on" is a recipe for getting bad
measurements and/or frying lots of nice, expensive equipment.
** When trouble shooting electronics, probes are often flying thick and fast
and mistakes are bound to be made.
The arrival of affordable, auto-ranging DMMs in the 80s was a real boon -
particularly so as the ohms ranges were immune from damage up to 240VAC on
most. One gets tired of replacing the low ohms resistor in an analogue
multimeter every few weeks.
But one thing I have learnt to be very careful with is *current*
easurement - forgetting to put the red probe back in the correct hole for
volts, ohms etc is a recipe for disaster.
The Fluke 70 series 2 that I mainly use has no current ranges and this is a
good thing !!
Also, wherever possible, I find a resistor ( or even a length of wire) and
measure the voltage drop across it on a DMM that can resolve 0.1mV DC or AC.
..... Phil
Phil Allison
Guest
Tue Jan 03, 2012 2:30 am
"George Herold"
"Phil Allison"
Quote:
** After just a little research - it is not hard to see why connecting a
scope directly to the AC mains is such a bad idea.
The vast majority of analogue and digital scopes have switched, passive
input attenuators using small resistors, trim capacitors and wafer
switches
employing several banks. All these components have max voltage ratings
that
do not include supply voltages like 240 volts AC.
Possibly the most vulnerable component to high voltages is the tiny trim
capacitor connected * directly across the input * to equalise capacitance
loading for all ranges - it is crucial that a scope's input present a
constant value of C ( usually 15 to 20 pF) in parallel with 1 Mohms on all
ranges so that the trimmer on a 10:1 probe can be adjusted once and left.
Wafer switches are also not rated for high voltages and lack sufficient
insulation and clearances to survive the spike voltages that regularly
appear on the AC supply. Insulation breakdown in the trimmer or switch
would be sudden and bit spectacular when full AC mains power is available.
Many scopes have a max input rating of 400 volts ( DC+AC) peak - which
means they are capable of accepting such peaks on a wave being viewed
without damage, despite the fact the actual peaks may not be shown. It
also
means that when AC coupled, the max DC voltage should not exceed 400 volts
as that is the rating of the series input capacitor.
The worst case of ACCEPTABLE input overload is with 400 volts at the input
and the attenuator set to pass the voltage straight through to the FET
preamp. Makers typically fit a series resistor of 100kohms to 470kohms
between the switch and the gate of the first FET to limit current to a few
mA - then add one or more diodes to direct this current safely into the
+/- DC supply rails. These diodes must be low leakage types so not to
create
an DC offset at the gate of the FET.
Cool, do they use the CB junction of a transistor?
** Not low leakage enough and low breakdown voltage.
Something like this is the go:
http://www.centralsemi.com/PDFs/products/1n3595.pdf
Max 1nA at room temp and 125V.
Some scope makers use the gate junction of a JFET - cos that has extremely
low leakage.
(none of the x10 probes I use have a switch! I hate that thing.)
** I hate swapping probes...
.... Phil
George Herold
Guest
Tue Jan 03, 2012 3:20 am
On Dec 29 2011, 8:06 pm, "Phil Allison" <phi...@tpg.com.au> wrote:
Quote:
** After just a little research - it is not hard to see why connecting a
scope directly to the AC mains is such a bad idea.
The vast majority of analogue and digital scopes have switched, passive
input attenuators using small resistors, trim capacitors and wafer switches
employing several banks. All these components have max voltage ratings that
do not include supply voltages like 240 volts AC.
Possibly the most vulnerable component to high voltages is the tiny trim
capacitor connected * directly across the input * to equalise capacitance
loading for all ranges - it is crucial that a scope's input present a
constant value of C ( usually 15 to 20 pF) in parallel with 1 Mohms on all
ranges so that the trimmer on a 10:1 probe can be adjusted once and left.
Wafer switches are also not rated for high voltages and lack sufficient
insulation and clearances to survive the spike voltages that regularly
appear on the AC supply. Insulation breakdown in the trimmer or switch
would be sudden and bit spectacular when full AC mains power is available..
Many scopes have a max input rating of 400 volts ( DC+AC) peak - which
means they are capable of accepting such peaks on a wave being viewed
without damage, despite the fact the actual peaks may not be shown. It also
means that when AC coupled, the max DC voltage should not exceed 400 volts
as that is the rating of the series input capacitor.
The worst case of ACCEPTABLE input overload is with 400 volts at the input
and the attenuator set to pass the voltage straight through to the FET
preamp. Makers typically fit a series resistor of 100kohms to 470kohms
between the switch and the gate of the first FET to limit current to a few
mA - then add one or more diodes to direct this current safely into the
+/- DC supply rails. These diodes must be low leakage types so not to create
an DC offset at the gate of the FET.
Cool, do they use the CB junction of a transistor?
Thanks Phil,
George H.
(none of the x10 probes I use have a switch! I hate that thing.)
Quote:
This high value series resistor needs a cap in parallel to pass high
frequencies that would otherwise be lost due stray capacitance to ground at
the gate of the FET - typically a 10nF cap is used and this too must be
rated for at least 400 volts.
In short - whenever scoping the AC supply make damn sure to use a 10:1
probe and be DAMN certain the switch is the right position.
... Phil
Phil Allison
Guest
Tue Jan 03, 2012 10:30 am
"fungus"
My reasoning was that a 'scope *has* to
have a very high impedance input. You don't
want it sucking current out of delicate little
circuits. Even at 10k ohms (quite low) you'll
only get 24mA from 240V A/C mains and
that's not enough to fry much.
** 240VAC across 10kohms is nearly 6 watts !!!
Enough to fry plenty.
Get real.
...... Phil
fungus
Guest
Tue Jan 03, 2012 11:11 am
On Jan 2, 9:43 pm, Bob Myers <bobmyer...@gmail.com> wrote:
Quote:
That was good advice in any event; you should ALWAYS
be in the habit of spending a good deal more time setting
up the measurement than actually making it. "Grab a
probe and stick it on" is a recipe for getting bad
measurements and/or frying lots of nice, expensive
equipment.
My reasoning was that a 'scope *has* to
have a very high impedance input. You don't
want it sucking current out of delicate little
circuits. Even at 10k ohms (quite low) you'll
only get 24mA from 240V A/C mains and
that's not enough to fry much.
But ... most 'scopes have some other
components before you get to the 10k,
I hadn't counted on that.
fungus
Guest
Tue Jan 03, 2012 11:16 am
On Jan 3, 12:23 am, "Phil Allison" <phi...@tpg.com.au> wrote:
Quote:
But one thing I have learnt to be very careful with is *current*
easurement - forgetting to put the red probe back in the correct hole for
volts, ohms etc is a recipe for disaster.
Yep. That one still catches me occasionally.
If I was richer (or electronics was more than a
hobby) I'd have two devices - one for current
and one for everything else.
fungus
Guest
Tue Jan 03, 2012 5:02 pm
On Jan 3, 10:30 am, "Phil Allison" <phi...@tpg.com.au> wrote:
Quote:
** 240VAC across 10kohms is nearly 6 watts !!!
Enough to fry plenty.
Get real.
Duh, I hadn't calculated the watts...
fungus
Guest
Tue Jan 03, 2012 5:35 pm
On Jan 3, 4:02 pm, fungus <to...@artlum.com> wrote:
Quote:
On Jan 3, 10:30 am, "Phil Allison" <phi...@tpg.com.au> wrote:
** 240VAC across 10kohms is nearly 6 watts !!!
Enough to fry plenty.
Get real.
Duh, I hadn't calculated the watts...
....but the '10k ohms' was just my bad example.
What I orignally meant to say was: I'd expect
the current a 'scope sucks out of a circuit to
be tiny, in the microamps range. Certainly less
then 1mA. Anything more could be enough to
give bad measurements.
A quick google says that typical 'scopes have
an input impedance of 1 megaohm so I was
in the right ballpark.
At one megahom the power from 240V A/C is
only 0.06 watts, which wouldn't fry much.
Fred Abse
Guest
Tue Jan 03, 2012 6:37 pm
On Mon, 02 Jan 2012 17:20:31 -0800, George Herold wrote:
Quote:
(none of the x10 probes I use have a switch! I hate that thing.)
Me neither. How many times do you need a x1 probe?
After enduring the fragility of Tek probes for years, I settled on German
made PMK ones. 500MHz with readout switching, and a slew of nice
accessories included. Better than half the price of a comparable Tek probe.
--
"For a successful technology, reality must take precedence
over public relations, for nature cannot be fooled."
(Richard Feynman)
George Herold
Guest
Tue Jan 03, 2012 7:33 pm
On Jan 2, 8:30 pm, "Phil Allison" <phi...@tpg.com.au> wrote:
Quote:
"George Herold"
"Phil Allison"
** After just a little research - it is not hard to see why connecting a
scope directly to the AC mains is such a bad idea.
The vast majority of analogue and digital scopes have switched, passive
input attenuators using small resistors, trim capacitors and wafer
switches
employing several banks. All these components have max voltage ratings
that
do not include supply voltages like 240 volts AC.
Possibly the most vulnerable component to high voltages is the tiny trim
capacitor connected * directly across the input * to equalise capacitance
loading for all ranges - it is crucial that a scope's input present a
constant value of C ( usually 15 to 20 pF) in parallel with 1 Mohms on all
ranges so that the trimmer on a 10:1 probe can be adjusted once and left.
Wafer switches are also not rated for high voltages and lack sufficient
insulation and clearances to survive the spike voltages that regularly
appear on the AC supply. Insulation breakdown in the trimmer or switch
would be sudden and bit spectacular when full AC mains power is available.
Many scopes have a max input rating of 400 volts ( DC+AC) peak - which
means they are capable of accepting such peaks on a wave being viewed
without damage, despite the fact the actual peaks may not be shown. It
also
means that when AC coupled, the max DC voltage should not exceed 400 volts
as that is the rating of the series input capacitor.
The worst case of ACCEPTABLE input overload is with 400 volts at the input
and the attenuator set to pass the voltage straight through to the FET
preamp. Makers typically fit a series resistor of 100kohms to 470kohms
between the switch and the gate of the first FET to limit current to a few
mA - then add one or more diodes to direct this current safely into the
+/- DC supply rails. These diodes must be low leakage types so not to
create
an DC offset at the gate of the FET.
Cool, do they use the CB junction of a transistor?
** Not low leakage enough and low breakdown voltage.
Something like this is the go:
http://www.centralsemi.com/PDFs/products/1n3595.pdf
Max 1nA at room temp and 125V.
Some scope makers use the gate junction of a JFET - cos that has extremely
low leakage.
(none of the x10 probes I use have a switch! I hate that thing.)
** I hate swapping probes...
... Phil- Hide quoted text -
- Show quoted text -
OK, I've heard that the CB leakage of transistors is spec'ed at ~10
nA. But if you measure a few it's much better. I got something like
10 pA for a 2n3904 at 10 volts. 'course the reverse voltage will not
be enough for a 'scope input.
George H.
krw@att.bizzzzzzzzzzzz
Guest
Wed Jan 04, 2012 1:00 am
On Tue, 3 Jan 2012 01:11:11 -0800 (PST), fungus <tooby_at_artlum.com> wrote:
Quote:
On Jan 2, 9:43 pm, Bob Myers <bobmyer...@gmail.com> wrote:
That was good advice in any event; you should ALWAYS
be in the habit of spending a good deal more time setting
up the measurement than actually making it. "Grab a
probe and stick it on" is a recipe for getting bad
measurements and/or frying lots of nice, expensive
equipment.
My reasoning was that a 'scope *has* to
have a very high impedance input. You don't
want it sucking current out of delicate little
circuits. Even at 10k ohms (quite low) you'll
only get 24mA from 240V A/C mains and
that's not enough to fry much.
Not all scopes have a high impedance. Many have a 50 ohm input impedance. You
don't want to be hooking them to the 240V mains.
Quote:
But ... most 'scopes have some other
components before you get to the 10k,
I hadn't counted on that.